Jose, CFSE being so bright on autoflourescence will make life very difficult for you to realize practical compensation values for your protocols. Try and optimize the PMT bias for CFSE alone and bring it as low as possible on FL1 Voltage gain. You will be happy with the other colors and most likely you'll get manageable compensations values before you exhaust your cell volumes to acquire. Thanks. Peter Mucheru. -----Original Message----- From: Jose Benito [mailto:jbenito1@hotmail.com] Sent: Monday, August 26, 2002 3:50 PM To: cyto-inbox Subject: CFSE analysis by flow Hi flowers, I am trying to work out a protocol for acquisition of cells labelled with CFSE and 3 more colors (PE, ECD and PE-Cy5). These are PBLs stimulated with mitogens and antigens. The problem id that CFSE gives a staining so bright (in the fourth log decade) that everything else (specially PE) is completely contaminated by the strong CFSE signal. I tried last friday to compensate for PE channel without succes (I run out of cells to acquire). I am going to try again today. Does anyone have any experience working with CFSE and other colors? Is tehre anyway to find out a cytosettings for this combination? Many thanks Jose Miguel Benito Madrid Spain _____ Join the world's largest e-mail service with MSN Hotmail. Click Here <http://g.msn.com/1HM1ENXX/c157??PI=31901>