Re: CFSE analysis by flow

From: J.H.N. Schuitemaker (j.h.schuitemaker@AMC.UVA.NL)
Date: Tue Aug 27 2002 - 04:53:41 EST


Dear Jose,

I'm very interested in your problem. From your mail I make up that you
work on a Coulter machine. We did the staining in quadriple (CFSE, PE,
PerCP and APC) without a problem on a BD machine. You can find an
example on http://home.wanadoo.nl/flowcytometry under CFSE. It is
possible to use CFSE and PE together. It might be a pain in the ....
But try to get make sample to set your machine. Use bulk PBMC, label and
stimulate them with SEB or PHA. Evaluate them after 4 to 5 days and you
will have your settings for the next experiments. At least you will be
close to them. We use 5 uM CFSE to stain them for 15 to 20 minutes. When
your signal is too strong the compensation will be more difficult.
Success and when you need any other suggestions....

Joost Schuitemaker


__________________________________________

J.H.N.Schuitemaker
Senior research technician
Cellular Immunology Group
Dept. Cell Biology & Histology
Academic Medical Center
P.O.box 22700
1100 DE Amsterdam
The Netherlands

Tel. + 31 20 5664960
Fax. + 31 20

----- Original Message -----
From: Jose Benito <jbenito1@hotmail.com>
Date: Monday, August 26, 2002 2:49 pm
Subject: CFSE analysis by flow

>
> <html><div style='background-color:'><DIV>Hi flowers, I am trying
> to work out a
> protocol for acquisition of cells labelled with CFSE and 3 more
> colors (PE, ECD and
> PE-Cy5). These are PBLs stimulated with mitogens and antigens. The
> problem id that CFSE
> gives a staining so bright (in the fourth log decade) that
> everything else (specially PE)
> is completely contaminated by the strong CFSE signal. I tried last
> friday to compensate
> for PE channel without succes (I run out of cells to acquire). I
> am going to try again
> today. Does anyone have any experience working with CFSE and other
> colors? Is tehre
> anyway to find out a cytosettings for this combination?</DIV>
> <DIV>Many thanks</DIV>
> <DIV>Jose Miguel Benito</DIV>
> <DIV>Madrid</DIV>