Dear Andrea, Some clones are better in a brighter conjugate. CD10 and CD34 are two of these. We use them in combinations where they are conjugated with PE or APC. For example CD19 FITC/CD10 PE/CD45 Percp / CD34 APC works well as does Kappa/Lambda/ CD19 Percp Cy5.5/CD10 APC. Neither is very good in FITC, no matter the vendor. CD10 in follicular lymphoma may also be dim which we have seen, so a brighter conjugate also makes sense. False positive CD34 results are trickier. Most good sales people will tell you if a particular clone is not good, but only if their customers ask. You might also consider an extra blocking step with mouse serum to get rid of FC receptors, use of 7-AAD or PI to detect dead cells which might non-specifically take up some stains, blocking with unconjugated CD34 to see if the positive staining disappears(equals real positive) and an IgG control which I consider vital when not looking as mature lymphocytes, as non-lymphs have more FC receptors. Hope this helps, Paula >Dear flow group, just a couple of questions for the experts.... > >CD10: We are currently using Caltag's CD10-FITC (clone 5-1B4) and seem >to have problem with staining on some of our follicular lymphomas. >Recently we had a case which was CD10-neg by flow but stained positive >with Ventanna's IPOX CD10 (clone 56C6). It seems to be Caltag's only >clone, Coulter has 3 clones (J5 which Coulter recommends, ALB1 and ALB2 >from the Coulter/ Immunotech merger). Since everybody uses a diggerent >terminology for their clones, can somebody tell me how they compare and >which clone is the most wiedely used by the flow community? > >CD34: We seem to have non-specific staining of CD34-FITC (Caltag's >clone 581 class III) on monos (we have a gate set around all the >populations and have noticed this fairly frequently). In the past we >have ignored this and we able to differentiate this from the real >specific staining. However, we had a recent case of LPD/B with ?CLL/?HCL >differentiation, where the monoclonal B-cells showed CD34 positivity >(TDT-neg, CD20+, CD22+, bright lambda+). Because of the mature phenotype >I was leaning towards ignoring it, but felt uneasy about totally >ignoring it. I just happened to have the red CD34 from Coulter (same >clone 581 class III) and those cells were negative! Has anybody else >seen this? > >Much appreciated! >Andrea Illingworth >Dahl-Chase Flow Cytometry >Bangor, Maine >207-990-4855 Paula Imus Flow Cytometry LP, NCI, NIH 10 Center Drive MSC-1500 Bethesda, Md. 20892-1500
This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:20 EST