RE: Detection of Nanoparticles

From: Nebe-Von-Caron, G (g.nebe-von-caron@unipath.com)
Date: Thu Aug 15 2002 - 07:11:48 EST


We measure latex particles down to 190nm with our Elite on 10 to 15 mW
aircooled laser power. Below 300nm we use side scatter triggering as
even our modified forward scatter cuts of at 300nm. We filter the sheath
"on line" in front of the flow cell. We could probably get below that by
cleaning up the sheath further as the scatter background signal goes
down if we stop the sheath flow. If I remember right from Howard the
theoretical limit is around 80nm but I don't remember at what
wavelength.

Regarding the display you have to look at two parameter plots of forward
versus side scatter as it allows you to remove some of the noise you get
for example from protein precipitates micro-airbubbles and other things
starting to interfere at this level.

Main secret seems to be a very clean sheath and flow cell and a good
alignment of the system.

Regards

Gerhard

-----Original Message-----
From: Ryan Duggan [mailto:rcduggan@midway.uchicago.edu]
Sent: 06 August 2002 18:10
To: cyto-inbox
Subject: Detection of Nanoparticles




Hello All,

Is it possible to detect particles on the order of 0.1um to 0.3um using
standard flow cytometers?  I've done some preliminary trials, and
haven't
had much luck (too much background noise, not enough FSC signal).  If
anyone has had any experience with this size particles, I'd appreciate
hearing from you.  The instruments I have to use are; FACScan,
FACSCalibur,
LSR, FACStar Plus, and MoFlo.  Also, If there are feasible modifications
I
can make to any of these instruments to be able to see this size
particles,
that would be helpful.	All ideas are welcome.

I've search the archives and found a few similar questions, but no
replies.

Thank You,
Ryan


_____________________________________________________________________

Disclaimer

This email and any files transmitted with it, is confidential and intended solely for the
use of the individual to whom it was addressed.

Unipath monitors emails to ensure its' systems operate effectively and to minimise
the risk of
viruses.  Whilst it has taken reasonable steps to scan this email, it does not accept
liability
for any virus that may be contained in it.


_____________________________________________________________________



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:19 EST