RE: EVC 304 abort rate

From: Mostowski, Howard S. (mostowski@cber.fda.gov)
Date: Mon May 13 2002 - 13:34:44 EST


David,

	The problem is not you, but the cells themselves, that is not
exactly accurate neither.  The cell are in a messy "FACS" media.  This
usually happens with any of the adherent cell and specially with cells that
are fixed.  I too have seen this problem appear, but it is around 50%  at 10
to 12k cells/sec at a sheath pressure of 14.  You probably noticed two other
facts [pardon the pun], but if you reduce the your sample pressure thus
reducing the number of cells/sec your instrument has to make a decision, the
abort rate drops dramatically. You did not mention if you had the Accu-drop
attachment, but if so you would notice that it is virtually impossible to
hold your side streams perfect even though they where with the control
beads. I use a two droplet-delay setting.
	Great now you know at least two of us has seen this phenomenon; that
does not explain what is happening and what I can do to somewhat alleviate
it.  So here goes.
	1. Filtering is a must as it will remove a great deal of the
artifacts cast off from the damaged cells as well as un-dissolved
preservative. You probably will not get rid of all the cellular debris as
more garbage is constantly being produce by the damaged cells sitting in
your sample tubes.  You might think you could reduce the problem by
increasing the threshold [I keep my FCS threshold at 20], it will not work.
The reason is, even though the artifacts are out of the picture
electronically they are still there physically.  Because they are still in
the equation, your instrument has to make a decision about cell + debris and
it is not in your favor.
	2. The other way to reduce the abort rate is as mentioned above is
to reduce your sample pressure to below 3000 cells/sec.
	3. The third way is to increase the concentration of cells. By
increasing the concentration of cell incorporated with a reduction of
cell/sec allows you to have a sample stream of a smaller diameter; thus
reducing the chance of debris coming past you beam at the same time, but
here you have to have a feather touch on your on your sample differential
knob.
	I hope this helps somewhat, Cheers and good luck.  If any ones else
has some other techniques, please put it in the open format, because I am
always ready to learn frame someone more wiser.

								Howard
-----Original Message-----
From: Corry, David [mailto:David.Corry@uwe.ac.uk]
Sent: Monday, May 13, 2002 7:23 AM
To: cyto-inbox
Subject: EVC 304 abort rate



Hi all,

I've recently started analysing ECV304 endothelial cells on my FACSVantage
and
have been having problems with a high co-incidence abort rate - typically
75%+!!
The cells are fixed with 1% paraformaldehyde/1% FBS in PBS and stained with
FITC or RPE conjugates. The problem doesn't seem to be as bad with unfixed
cells or with some of the other adherent cell lines I use e.g. C20/A4
chondrocytes.
Could it be a problem with the fixing protocol? Anyone got any ideas or
suggestions?

Thanks in advance

Dave

----------------------------------------
David Corry
Centre for Research in Biomedicine