Dr. Shank,

Bangs Laboratories offers a product line called Quantum™ MESF Kits that would work well
for you.  Each consists of 5 bead populations- 1 blank and 4 fluorescent populations
labeled with varying intensities of a specific fluorochrome.  Each of the 4 stained beads
is assigned an MESF value (Molecules of Equivalent Soluble Fluorochrome).  When you run
them on the cytometer, you'll get 5 separate peaks of fluorescence.  These are used to
set up a standard curve based on the mean channel value of each peak in your histogram.
The R-squared value (regression) of that line is your check of linearity in that channel.
The kits are available labeled with FITC, PE, and PE-Cy5.

In the interest of not using the message board as a commercial, I would also have you
know that Molecular Probes offers a similar kit called "LinearFlow™" and Spherotech
offers one called "Sphero™ Rainbow Calibration Particles."

Each of the kits confers its own, unique advantages.  I would suggest that you contact
all three of us to determine which product may be the best choice for your application.

>>I have a B/D FACScan and two Coulter XLs.  I'm looking for the quintessential
multi-level
>>fluorescent bead product to do linearity and sensitivity, compatible with both Coulter
>>and B/D.

Best regards,

Nathan D. Foushee, MT (ASCP)
Product Manager - Flow Cytometry

Give us a call- Let's MultiPlex!
www.QuantumPlex.com

Bangs Laboratories / FCSC
9025 Technology Drive
Fishers, IN 46038-2886
800.387.0672     317.570.7020
fax 317.570.7034
nathan@bangslabs.com
www.bangslabs.com



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:34 EST