Re: FACScalibur instrument settings for PI ploidy/apoptosis acquisition

From: Nathan Regimbal (Nathan.Regimbal@menlo.ppdi.com)
Date: Thu Aug 02 2001 - 11:04:06 EST


Hi all,

How can one be sure that the sub G1 events are apoptotic events, and not just
DNA fragment clusters of varying sizes?  Wouldn't marker statistics of the
sub-G1 'pile' also be inaccurate, as the marked events are not actually cells,
but fragments of DNA?  Also, regarding the presence of a sub-G1 'apoptotic'
peak: what would cause apoptotic cells to all display slightly lower, yet
Gaussian, DNA content?   Doesn't a sub-G1 peak mean that these cells all lost
the same, small amount of DNA?  My apologies if this has been covered in the
past.

Cheers,
Nate Regimbal
PPD Discovery
Menlo Park, CA

Tim Kute wrote:

> In most cases, I like to have my G1 peak start  in the 1/3 rd of the x axis
> range.  If you are looking for apoptotic cells, you might want it higher
> around 2/3 rd of range.  If you really want to see every event, you might
> even want to put the scale on log and have the G1 peak at the 80% area of
> the x axis.
>
> Tim Kute
>
> Maciej Simm wrote:
>
> > I'm posting this on behalf of a collegue. She's doing an assay with
> > just PI looking at the peaks before/after incubation of her cells in
> > pro-apoptotic drugs.  If anyone looks at PI on a Calibur, could you
> > fill me in on the instrument settings that work the best? I would
> > appreciate any other tips.
> >
> > Maciej
> >
> > >find out voltage settings for the PI assay. in particular ask  about
> > >a printout of an acquisition template they use. also -
> >
> > >fsc    lin or log? what voltage?
> > >ssc   same as above
> > >fsc    s.a.a.
> > >fl2     s.a.a.
> > >fl3     s.a.a.
> >
> > >what about compensation?
> >
> > __________________________________________________
> > Do You Yahoo!?
> > Make international calls for as low as $.04/minute with Yahoo! Messenger
> > http://phonecard.yahoo.com/

______________________________________________________________________
This email transmission and any documents, files or previous email
messages attached to it may contain information that is confidential or
legally privileged. If you are not the intended recipient or a person
responsible for delivering this transmission to the intended recipient,
you are hereby notified that you must not read this transmission and
that any disclosure, copying, printing, distribution or use of this
transmission is strictly prohibited. If you have received this
transmission in error, please immediately notify the sender by telephone
or return email and delete the original transmission and its attachments
without reading or saving in any manner.
______________________________________________________________________



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:27 EST