Re: Cells Annexin-V and PI +

From: Derek Davies (daviesd2@icrf.icnet.uk)
Date: Tue May 29 2001 - 03:54:22 EST


Hi Camila,

On a dual plot of annexin (x axis) v PI (y axis), your live cells will
be in the bottom left corner. Early apoptotic cells will be in the lower
right quadrant ie annexin+ but PI-. As cells become more permeable to PI
they will appear in the upper right quadrant ie will be Annexin+ and
PI+. These cells will be dead ie late apoptotic and necrotic. Later on
if they lose cytoplasm, shrink, generally blow up there may be a
population that is annexin- but PI+. A quick look down the fluorescence
microscope will help identifying these.

You have to remember that in all live cells assays like this, we are
looking at a snapshot of a dynmamic process. If you have induced
apoptosis cells will be cointinually moving in an anticlockwise loop on
that dual parameter plot starting in the lower left quadrant. Also you
may find that some cell types give a nice clear annexin only population
whereas others will have an 'apoptotic' population that is also slightly
PI positive presumably due to slight permeability cahnges in early
apoptosis.

Derek


On Thu, 24 May 2001, Camila Rodrigues Cacere wrote:

>  We have been studying the levels of apoptosis in peripheral blood
> mononuclear cells of patients with paracoccidioidomycosis. These cells
> are cultivated for 4-7 days with a specific fungal antigen, and we are
> assessing the apoptosis through Annexin-V and PI.
> Do someone know how to interpret those cells  that are double stained
> (FITC+PI): cells that are in late apotpsosis process? Necrosis? In our
> cultures we detect very few cells stained with PI only, but significant
> numbers of cells stained with FITC only, double stained cells, or (the
> majority, live) non-stained cells.
>

************************************************************************
Derek Davies                       Voice: (44) 020 7269 3394
FACS Laboratory,                   FAX: (44) 020 7269 3100
Imperial Cancer Research Fund,     e_mail: derek.davies@icrf.icnet.uk
London, UK			   mobile: 07790 604112

Web Page: http://www.icnet.uk/axp/facs/davies/index.html

In tenebris lux
*************************************************************************



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:20 EST