What beads are you using? I'm guessing you don't add PerCP if you use the BD beads, 'cause the PerCPs are very unstable in my hands. If I mix all 4 (unlabeled, FITC,PE, PerCP) they are useless after about an hour. (After that I no longer have single-color beads!) I have to make new dilutions each time. Do you know something I don't? David McFarland Howard Hughes Medical Institute Flow Cytometry Facility Vanderbilt University Medical Center "Jane S. Miller" <miller@medicine.tamu.edu> on 03/30/2000 01:23:34 PM (Embedded image moved to file: pic08453.pcx)From:(Embedded image moved to file: pic05728.pcx)"Jane S. Miller" <miller@medicine.tamu.edu> on 03/30/2000 01:23 PM To: Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu> cc: (bcc: David McFarland/VUMC/Vanderbilt) Subject: Re: FACSComp QC I run FACSComp each day that I run immunofluorescent samples. A mix of beads will keep a week or more. I don't make a new mix every time. Jane Miller >>> Tami Rosario <trosario@cellmate.cb.uga.edu> 03/29/00 12:55PM >>> I have a question for those of you who run FACSComp on the FACSCaliber for QC. Do you run this each time you start up the cytometer or only occasionally. I have been told that I should run FACSComp each day that I run samples, but the beads are fairly expensive. Any alternatives? Thanks, Tami