RE: FACSComp QC

From: Abby Kelliher (allena@helix.mgh.harvard.edu)
Date: Mon Apr 03 2000 - 15:51:37 EST


We are a clinical lab and we started running FACSComp once per day, but
found it to be unnecessary.  We were able to document that our voltages did
not change much at all from week to week, so we now run FACSComp once per
week, unless 1) the instrument is cleaned, 2) it is serviced, 3) there is a
noticeable problem.  It is good for calibrating the instrument, but most
problems I see are not diagnosed by FACSComp, and I don't recommend relying
on it for that purpose.


Abby Kelliher
Clinical Flow Cytometry Laboratory
Warren Bldg.  Room 112
Massachusetts General Hospital
275 Charles St.
Boston, MA  02114

617-726-8487


-----Original Message-----
From:	CJett [SMTP:CJett@Compucyte.com]
Sent:	Friday, March 31, 2000 12:49 PM
To:	Cytometry Mailing List
Subject:	RE: FACSComp QC


CJ> We were originally running FACSComp at the beginning of each day...but
later found our stability was good enough to go to a weekly check
up.....but
you have to be vigilant for voltage creep on your laser..once you get
around
3000 hrs the 488nm argon can start to drop off and get noisy.  Of course a
clinical lab will not want to do this..once a day, keeps the FDA away



>>> Tami Rosario <trosario@cellmate.cb.uga.edu> 03/29/00 12:55PM >>>

I have a question for those of you who run FACSComp on the FACSCaliber
for QC.  Do you run this each time you start up the cytometer or only
occasionally.  I have been told that I should run FACSComp each day that
I run samples, but the beads are fairly expensive.  Any alternatives?