We are quantitating cell associated antigens using standardized beads and then generating an MESF from the standard curve generated from the beads. The question has come up whether or not to "correct" the MESF values obtained from the specific antibody by subtracting the MESF values obtained for the isotype matched controls. The problem with correcting the values is that it has little effect on the samples with high density of Ag, but has a large effect on those with low expression. As such, if we run replicate samples on the low expressing samples we find a larger amount of variation in the corrected MESFs. My bias is not to use it, as it seems to introduce more "noise" into the system. Your thoughts? Thanks! > _______________________ > Calman Prussin > Laboratory of Allergic Diseases > NIAID/ National Institutes of Health >
This archive was generated by hypermail 2b29 : Sat Mar 10 2001 - 19:31:02 EST