From: David_C_McFarland@sbphrd.com
Date: Tue Jun 27 2000 - 16:05:43 EST


Hi Warren.  It's the APC beads.  I don't know what the deal is but they crap out
in a hurry.  I think the normal target value for the APC beads is about 212.  Go
into the FACSComp setup and change the target value to about 175.  You were
receiving that error message because you aren't getting enough signal out of the
APC beads for them to be recognized.  You could get some new APC beads, however,
they will probably lose intensity quickly so you might as well know how to do
this anyway.  If these suggestions don't do the trick it could actually be the
laser.  I had one fail after about 3 years.

David McFarland
SmithKline Beecham Pharmaceuticals


---------------------- Forwarded by David C McFarland/DEV/PHRD/SB_PLC on
06/27/2000 16:58 ---------------------------


wds8@email.med.yale.edu on 26-Jun-2000 17:07



To: cyto-inbox
cc:    (bcc: David C McFarland/DEV/PHRD/SB_PLC)
Subject:  problems with time delay calibration on FACS Calibur





Dear Flowers (in particular FACS Calibur users),

We have a new FACS Calibur and have had persistent problems with
reproducibly being able to set time delay calibration.  In
particular, we get the error message, "signal out of range...unable
to set time delay."  BD has replaced an EPROM and a board, but the
problem intermittently persists.  In speaking to users of 3 other
Caliburs I discovered that they too had seen this problem.  I'd
appreciate any comments from other two laser users on the Calibur and
will summarize the responses to the list.  Many thanks.


Warren