Re: problems with time delay calibration on FACS Calibur

From: Simon Monard (smonard@trudeauinstitute.org)
Date: Wed Jun 28 2000 - 08:25:45 EST


Hi

The sheath pressure regulator doen't seem to very good on Caliburs, the pressure will
vary a bit each time you pressurize the tank. The time taken for a particle to pass
from the red to blue laser is dependant on sheath pressure. Its worth investing in a
pressure gauge so you can check it. You should also check that the sheath tank cap is
on tightly and is not leaking. It is pretty easy to  tweak the laser but perhaps you
should get your engineer to show you how to do it. I insist that folk do the time delay
calibration each time the tank is filled, I use Molecular Probes 6u alignflow beads.

Good luck

Simon Monard
FACS Lab Manager
Trudeau Institute
Saranac Lake
NY12983

Ph 518 891 3080 X352



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:55:54 EST