RE: CLL PHENOTYPE

From: Van Bockstaele, Dirk (Dirk.Van.Bockstaele@uza.uia.ac.be)
Date: Thu Jul 15 1999 - 05:32:07 EST


Hi Sebastian,
Never use anti-kappa, lambda reagents on whole blood!!!  The ab binds to the
serum Ig that is abundantly present, so there's nothing left over to bind on
the B-cells surface Ig.  Not long in the business I guess?
Best Regards,
Dirk

Prof. Dirk Van Bockstaele, PhD
Laboratory of Hematology
Antwerp University Hospital
Wilrijkstraat 10
B-2650 Edegem
Belgium
tel. 32 3 821 3900, fax 32 3 825 1148, 
e-mail dirk.van.bockstaele@uza.uia.ac.be


> ----------
> Van: 	Sebastian, J, Gelderbloem, Mr[SMTP:sjg@samiot.uct.ac.za]
> Verzonden: 	woensdag 14 juli 1999 19:09
> Aan: 	Cytometry Mailing List
> Onderwerp: 	Re: CLL PHENOTYPE
> 
> 
> Dear Flowers
> 
> I have a slight problem and I am not sure what the solution is.
> 
> We use polyclonal Kappa and Lambda from BD for our chronic 
> leukaemia panel. We have been setting the samples up using whole 
> blood and for some or other reason the Light Chains are coming up 
> negative everytime. We only bitmap lymphoid cells.
> 
> Is there anyone with some advice.
> 
> Thanks 
> Sebastian
> 