Re: CLL PHENOTYPE

From: Anna Porwit-MacDonald (Anja.Porwit@mb.ks.se)
Date: Thu Jul 15 1999 - 04:40:38 EST


Hi,
the expression of light chain on CLL is very low. We use also the whole
blood assay but usually wash the cells before staining to remove serum. If
your assay is well standardized in terms of the same fluorescence intensity
for normal kappa and lambda cells, you can compare fluorescence intensity
of kappa and lambda by Kolmogorov-Smirnov analysis.
Best wishes
Anna

At 15:09 1999-07-14 UTC-2, you wrote:
>
>Dear Flowers
>
>I have a slight problem and I am not sure what the solution is.
>
>We use polyclonal Kappa and Lambda from BD for our chronic 
>leukaemia panel. We have been setting the samples up using whole 
>blood and for some or other reason the Light Chains are coming up 
>negative everytime. We only bitmap lymphoid cells.
>
>Is there anyone with some advice.
>
>Thanks 
>Sebastian
>
>
Anna Porwit-MacDonald
Heamatopathology Lab.
Department of Pathology
Karolinska Hospital, Stockholm
anpo@mb.ks.se
tel.:+46-851775863
fax.:+46-851775843



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:53:43 EST