Re: Ratiometric calcium measurement

From: Lynn Dustin (dustinlb@slu.edu)
Date: Fri May 07 1999 - 11:51:16 EST


Beth,
We use a 3-color FacsCalibur with a 488 nm laser.  I use Fluo3/Fura Red or
(better) Fluo4/Fura Red to measure calcium mobilization in B cells (J.
Immunol. 162:2717). I also helped another lab here use Fluo3/Fura Red in
human PMNs with very good results. Fluo3 alone is less sensitive and less
consistent than Fluo3+ Fura Red as a calcium indicator.  Computing the
ratio of fluorescence from 2 dyes allows you to correct for cell to cell
variability in dye loading.
Good luck
Lynn Dustin