Re[2]: GFP/Apoptosis

From: Tom Frey (Tom_Frey@bdis.com)
Date: Wed Jul 30 1997 - 11:04:25 EST


A few small additions to Mario's note.  First, annexin V biotin is also 
commercially available and indirect staining may be easier than trying to do a 
conjugate.  Second, in my experience annexin V staining will not survive PFA 
fixation unless some divalent cation (calcium) is present in the fixative (and 
any wash steps!).  I have only used bovine annexin V, but I assume that this 
will be true of human recombinant annexin V also.

Tom Frey


We haven't specifically tried GFP with annexin, but there is no reason for this
not to work--we often use GFP with a wide variety of fluorochromes, like PE,
Cy5PE, TR, APC, etc; annexin V can be conjugated to any of these to be used in
combination with GFP.  If you have the annexin V protein (it can be obtained
both commercially and in collaboration with some scientists producing
recombinant versions), conjugation instructions can be found at
"http://cmgm.stanford.edu/~roederer/abcon/".

BTW, the annexin V staining survives 0.5% paraformaldehyde fixation just fine.

mr


>   Has anyone tried this combination? Or are there any other
>   apoptosis  markers that would work with GFP?