>our question is whether using ionomycin (as our positive control/at various
> concentrations) should the ratio of violet to blue return to near baseline
> after the influx of calcium?
>
In our hands, working mainly with mouse thymocytes, the ratio stays high
after the addition of ionomycin until such a time as the cells begin to
become "less healthy" as indicated by a decreasing forward scatter, slightly
increased side scatter. If you let the time course continue you can plot
nicely the return to near baseline of indo-1 fluorescence ratio correlating
with scatter change in a 3D type plot.
As a rough method of duplicating this effect, try adding a squirt (technical
measurement I use) of 70% ethanol to the end of one of your sample time
courses and you should see a similar pattern to the above ionomycin
treatment. This doesn't hang around in the sample lines, and can be a quick
way of checking whether you've got any loading if ever you don't get a
response from your agonist.
Hope this helps,
Geoff
======================================================================