Re: CD34 cell enumerations
Margaret Cooley (m.cooley@arnie.cfi.unsw.EDU.AU)
Wed, 17 May 1995 09:02:25 +1100
>
>To:G R BARCLAY <100066.300@compuserve.com>
>From:m.cooley@cfi.unsw.edu.au (Margaret Cooley)
>Subject:Re: CD34 cell enumerations
>
>We routinely do CD34+ estimations on growth-factor mobilised peripheral blood
>leukapheresis specimens. The set up we use is CD45-PerCP from B-D, then
>CD34-FITC or CD34-PE (both HPCA-2 from B-D, but some labs use a cocktail of
>HPCA-2 and Qbend anti-CD34 from Immnotech, as these see different epitopes and
>the combination may give better results). This gives you a third colour to
>use if you want to look at subsets of CD34+ cells, e.g. CD33+ or CD38+. We
>just stain for 10 min room temp and treat the specimen on Coulter Q-prep, wash
>once afterwards.
>
> The machine set up is:
>
>Hist 1: CD45-PerCP as single parameter histogram, then set gate on all CD45POS
>cells (n.b., stem cells are relatively dim CD45 so be generous on the lower
>limit): this just excludes debris and any unlysed red cells.
>
>Hist 2 (gated on Hist 1): CD34 versus side scatter. This gives you a very
>clear population of low side scatter CD34-bright cells which represent the
>true stem cells, and means the dim CD34+ cells which are more mature and have
>limited repopulating potential are excluded.
>
>This set up enables you to calculate the CD34+ as a percent of total nucleated
>cells without trying to estimate % mononuclear from sometimes indeterminate
>scatter pictures.
>
>N.b. count a minimum of 100,000 nucleated cells, unless your percent CD34+ is
>over 1%.
>
>We use a Coulter EPICS XL but I see no reason why this strategy could not be
>used on any analyser or sorter.
>
>Bst of luck,
>
>
Margaret A. Cooley, PhD
Centre for Immunology
St Vincent's Hospital
Darlinghurst NSW 2010 AUSTRALIA
Ph: 61-2-3617700; Fax: 61-2-3612391.
CD ROM Vol 2 was produced by staff at the Purdue University Cytometry Laboratories
and distributed free of charge as an educational service to the cytometry community.
If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL,
Purdue University, West Lafayette, IN 47907. Phone:(317) 494-0757; FAX (317) 494-0517; Web http://www.cyto.purdue.edu
EMAIL robinson@flowcyt.cyto.purdue.edu