RE: Static Charge on Sort Collection Tube

From: Nebe-Von-Caron, G (g.nebe-von-caron@unipath.com)
Date: Thu Nov 28 2002 - 10:10:59 EST


Just to confirm that I do the same on the EPICS Elite by using the sort
stream centring pot. You do indeed loose the right or left sort stream
depending of the way you turn it but are left with the uncharged drop.
As variations in droplet size and charge are effecting the precision of
the deflection movement you can deposit much more accurately.

The other thing I try to avoid when sorting bacteria is to sort directly
into a liquid surface. This leads to the release of aerosols that, if
charged, can nicely jump between sort vessels. For longer sorts I use a
cleaning rod from an old Coulter counter, very thin old steel string,
cleaned in isopropanol.

Regards

Gerhard

-----Original Message-----
From: Mario Roederer [mailto:roederer@drmr.com]
Sent: 26 November 2002 20:35
To: cyto-inbox
Subject: Re: Static Charge on Sort Collection Tube



The cheap and easy solution (depending on your hardware) to avoid
charge accumulation for one-way sorting is to push the central waste
stream collection off to the side (Dick Stovel taught us to use a
chopped-off yellow tip of the right length, wedging it in to push the
collector over)--set the voltage balance to push the central stream
into the waste collection, and the sort stream to be absolutely
vertical (i.e., no net charge).  You can aim a drop into a very small
area using this method, and for large sorts, there's no charge
accumulation.

It would be nice if manufacturers would make the necessary
accomodations to make "vertical" sorts easier to set up; in my
opinion, there's no reason it shouldn't be done anytime you do a
one-way sort!

mr

At 10:35 PM -0500 11/25/02, Howard Shapiro wrote:
>Larry Polsky wrote-
>
>>As a newcomer to sorting, I would appreciate any thoughts or comments
on
>>the need for inserting a static drain wire into non-conducting
>>sort-collection tubes.
>>
>>There are references in the literature (see below*) that note the
>>possibility of sorted droplets being repelled by static charge that
can
>>accumulate on non-conducting, plastic collection tubes. The suggestion
>>made in these references is to insert a grounded drain wire into the
>>collection tube to dissipate any charge. However, two experienced
sorters
>>who I have asked about this have not seen static accumulation to be a
>>problem and do not use drain wires.  The suggestions have also been
made
>>to use glass tubes (not always possible) or to coat non-conducting
>>collection tubes with serum to achieve the same result.
>
>The droplets you sort are charged. If you sort them into a
nonconductive
>vessel, they stay charged (you can pick up a shock by dipping your
finger
>into a weight boat full of sorted droplets, but it's kind of hard to
get a
>finger into one of the small tubes typically used for sorting). Once
you
>accumulate a pool of charged droplets, they tend to repel droplets of
like
>charge. I don't sort much, but when I have done sorts, I have noted
some
>strange behavior of droplets, which seemed to go away when I used a
>grounding wire. If you are not inexplicably losing cells, you probably
>don't need to bother with grounding. If accumulated charge got to be a
>problem, the best solution might be to have collection tubes made of
>conductive plastic.
>
>-Howard