Re: GFP sorting question

From: Gisele Knowles (gknowles@sten.sunnybrook.utoronto.ca)
Date: Fri Oct 18 2002 - 08:53:34 EST


> Hi Andy,
> what are the gfp tagged cells?  if they are a substrate attached cell line,
> they are substantially larger than pblšs to sort.  You need to make bigger
> drops.  Try bumping up the nozzle size (that will involve slower speeds and
> only do it if you can get decent side stream separation) or as a starter try
> to sort on normal R mode with 1.0 drops and see if that cleans up the side
> streams with the cells running through.  With the cells running, also gently
> tweek the amplitude and that might also help clean it up.
> hope this helps and good luck
> 
> 
> Hi FLow-er in Flow land,
> I've come across an interesting (and confusing) problem which sorting GFP
> cells.  I've noticed several times that I get alot of fanning while sorting.
> I don't believe it is the Frequency, drive or phase since setting up the sort
> gives me beautiful side streams and great separation.  I first notice this on
> my Elite and now I have a FACSVantage and 2 of the last 3 GFP sorts showed
> this problem. After todays sort, which had terrible fanning, I put on a sample
> of unstained fixed PBL's to see if it was instrument related.  Not
> surprisingly, that sample gave great side streams.  Obviously the problem is
> with the sample.  I've never (never say never!) seen a fanning problem with
> surface stained cells and I don't understand what could be causing this to
> happen with GFP cells?  Could the transfection method somehow "charge" the
> cells which interferes with the charge applied for sorting?  As far as I can
> tell the media (which is MEM with Pen, strep, L-Glut, FCS, bicarb.) shouldn't
> do anything?
> 
> Any help or suggestions would be greatly appreciated!
> 
> Thanks in advance.
> 
> Andy
> 
> 
> ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^
>   Andy Oberyszyn, M.S.
>   Core lab manager
>  The Ohio State University
>  University Cell Analysis & Sorting Core
>   424 Davis Heart & Lung Institute
>   473 West 12th Avenue
>   Columbus, Ohio 43210
>   Tel: 614/292-FLOW(3569)
>   Fax: 614/292-7335
>   E-Mail: cytometry@osu.edu
>   Website: http://heartlung.osu.edu/hlri/corelabs/flowcore.jsp
>                  
> ^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^
> "What if the Hokey Pokey is really what it's all about?!?"
> 

-- 
Gisele Knowles
>>Director
Centre for Cytometry and Scanning Microscopy
Sunnybrook and Women's Research Institute
Room B 210
2075 Bayview Avenue
Toronto, ON
416.480.6100x7282 (office)
x7284/5 (lab)



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:27 EST