Molecular probes sells an anti-PentaHis antibody

http://www.probes.com/servlets/product?region=Select+Region&item=21315

that crossreacts with hexahistidine-tagged proteins in most applications. It is
a mouse IgG1 antibody, indicating that it will complex with our new Zenon One
reagents

http://www.probes.com/handbook/sections/0702.html

to form labeled antibodies (one can make the complex either before or after if
one uses a single color, except as below). We have had generally good success
using Zenon labeling complexes for intracellular staining of antigens by flow
cytometry (and imaging) although access may depend on your permeabilization
techniques.

However, if you also have cell surface expression of hexahistidine-tagged
proteins (or other proteins that crossreact with the same primary antibody), it
will be necessary to block these first while the cell is alive. We have never
tried it but presumably one could block the cell-surface-expressed
hexahistidine-tagged proteins with the same unlabeled mouse antibody to
pentahistidine then, after fixation and permeabilization, use the pre-made Zenon
One reagent-labeled complex of the same primary anti-pentahistidine antibody to
label only the intracellular proteins. One could not do this with a secondary
anti-mouse antibody because it would crossreact with the cell-surface-bound
primary antibody but it should be possible with the Zenon One complex, which is
essentially equivalent to the covalent dye conjugate of the mouse primary. The
Zenon One label will not transfer to the anti-pentahistidine antibody used as
the blocker, at least over the period of the experiment. The limitation would be
whether the unlabeled primary has complete access to the cell-surface-expressed
hexahistidine fusion proteins to block their binding activity step of the Zenon
One reagent-labeled primary (if that is what is giving you background). Could be
interesting to try this and I cannot think of another way to do the same
experiment.



Shelia R Alexander wrote:

> Hello
> I am searching for a 6His tag antibody that works by flow cytometry as a
> marker for transfection efficiency.   I have stably transfected K562 and
> U937 cells and a mouse fibroblast line and put them under puromycin
> antibiotic selection and have a population that has grown out but need to
> prove that they express my protein of interest and aren't antibiotic
> resistant.    What I have seen so far is that the 6-His monoclonal
> antibodies stain extracellular and permeabilized cells from my parent and
> transfected cell line equally well.    My 6-His tag is expressed
> intracellular in this case but future studies would place the His tag on the
> surface of the cells.   Are there ways to block "native" HIS expression?
>
> Any suggestions would be appreciated.  Thank you
>
> Shelia Alexander