Re: FRET frets...

From: Richard Konz (RKonz@wyeth.com)
Date: Tue Jul 02 2002 - 14:40:54 EST


Greetings again everyone:

FRET frets solved.

Set up unstained and FITC [or PE] and CY5 controls with the diode laser on and collect
CY5 signal in FL4.

After individual controls are run, turn off the diode laser [eliminates
FL4] and look for CY5 FRET in FL3 generated from the resonance energy of FITC or PE.
FL3 is a 670LP in the FACScalibur so CY5 looks great when you get
FRET from FITC or PE.

Thanks to Alice Givan for reminding me to check this out.

All my best,

-Rich

Richard F. Konz, Jr.
Scientist
Manager, Flow and Image Cytometry Facility
Wyeth
200 Cambridge Park Drive
Cambridge, MA 02140   USA
Office: 617-665-5522
Lab: 617-665-5551

>>> "Richard Konz" <RKonz@wyeth.com> 07/01/02 01:55PM >>>

Greetings everyone:

An investigator in my research group is doing FRET by facs analysis
on a two laser FACScalibur [FITC/PE and CY5].  During experimental setup
we use the red diode laser to look at the CY5 staining signal.  However when we do the
FRET experiment we need to have the diode laser off.  When you shut the diode laser off,
Cell Quest will eliminate FL4 as a collection parameter.

Thus far, we are unplugging the diode laser from the board while we are doing the FRET
experiment.  I could install a switch to cut the power, but in a multi-user facility
this would probably cause problems with the users forgetting to turn it back on.

Does anyone have any ideas or potential workarounds with this FRET combination?

Thanks for your input and all my best regards,

-Rich



Richard F. Konz, Jr.
Scientist
Manager, Flow and Image Cytometry Facility
Wyeth
200 Cambridge Park Drive
Cambridge, MA 02140   USA
Office: 617-665-5522
Lab: 617-665-5551



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:14 EST