Re: FRET frets...

From: Mario Roederer (roederer@drmr.com)
Date: Tue Jul 02 2002 - 14:27:07 EST


You don't have to have the laser off to measure FRET.  This
misconception arises probably from microscopy, where you don't have
temporally distinguished excitation....

Anyway, the FL3 signal will have your FRET signal irrespective of
whether or not the diode laser is on or off--so just leave the laser
on.

FRET signal will not appear in the FL4 channel anyway--it comes from
first laser excitation, therefore you wouldn't see it in the FL4
channel!

mr

At 1:55 PM -0400 7/1/02, Richard Konz wrote:
>Greetings everyone:
>
>An investigator in my research group is doing FRET by facs analysis
>on a two laser FACScalibur [FITC/PE and CY5].  During experimental setup
>we use the red diode laser to look at the CY5 staining signal.
>However when we do the
>FRET experiment we need to have the diode laser off.  When you shut
>the diode laser off,
>Cell Quest will eliminate FL4 as a collection parameter.
>
>Thus far, we are unplugging the diode laser from the board while we
>are doing the FRET
>experiment.  I could install a switch to cut the power, but in a
>multi-user facility
>this would probably cause problems with the users forgetting to turn
>it back on.
>
>Does anyone have any ideas or potential workarounds with this FRET
>combination?
>
>Thanks for your input and all my best regards,
>
>-Rich
>
>
>
>Richard F. Konz, Jr.
>Scientist
>Manager, Flow and Image Cytometry Facility
>Wyeth
>200 Cambridge Park Drive
>Cambridge, MA 02140   USA
>Office: 617-665-5522
>Lab: 617-665-5551



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:14 EST