Simona wrote: Dear Flow users, I want to check the NK cells in blood mouse. I used the antibody pan-NK cells DX5 ( Pharmingen) but I had 35% posite cells. So I think that this antibody is not specific only for NK population and I don't know what other populations are recognized. Can you help me? Thanks Bye Simona We have used the DX5 from Pharmingen on spleen cells from mice. Of total lymphocytes, it routinely showed only 3 - 5% NK cells. Have you titrated the DX5? We use 500 ng per million spleen cells and allow it to stain for 30 minutes on wet ice. What blocking agent(s) are you using? We use PBS with 0.2% BSA and 0.02% sodium azide. Delynn Moss Centers for Disease Control Division of Parasitic Diseases -----Original Message----- From: Simona Ronzoni [mailto:sronzoni@ieo.it] Sent: Monday, May 20, 2002 11:47 AM To: cyto-inbox Subject: NK cells Dear Flow users, I want to check the NK cells in blood mouse. I used the antibody pan-NK cells DX5 ( Pharmingen) but I had 35% posite cells. So I think that this antibody is not specific only for NK population and I don't know what other populations are recognized. Can you help me? Thanks Bye Simona Simona Ronzoni Dept. of Experimental Oncology European Institute of Oncology via Ripamonti, 435 20141 Milan Phone: +39 02 57489880 Fax: +39 02 57489851