From: Janos, Szollosi (szollo@jaguar.dote.hu)
Date: Mon May 20 2002 - 08:46:33 EST


Dear Richard!

Yes, we have used the FacsCalibur to perform FRET measurements using Cy3
and Cy5 as donor acceptor pair.  The 488 nm excitation is not optimal for
Cy3 (543 would be better), 632 is optimal for Cy5, and with this setup we
could obtain reasonable good FRET distribution curves (actually they are
better than that obtained with FITC/TRITC pair because the autofluorescence
is much less of a problem).  In addition we could use the 488->520 band for
autofluorescence correction on a cell-by-cell basis.  The method was
presented in the ISAC meeting in San Diego, and a manuscript has been
accepted in the CYTOMETRY and will be published within a few month.  If you
are interested we can send the manuscript.

Long Wavelength Fluorophores and Cell-by-Cell Correction for Autofluorescence
Significantly Improves the Accuracy of Flow Cytometric Energy Transfer
Measurements on a Dual-laser Benchtop Flow Cytometer

Zsolt Sebestyén1, Péter Nagy2, Gábor Horváth1, György Vámosi2, Reno
Debets3, Jan W. Gratama3, Denis R. Alexander4, János Szöllősi1

In addition we have elaborated a software which can calculate the FRET
efficiency on a cell-by-cell basis and made it available for the
public.  We have also included some help file and some sample data.

The web page is:
www.biophys.dote.hu/kutatas/research.htm and link to follow: Instruments,
software. The FLEX program is on the side.  We are still working on the
help instruction, and as we proceed we constantly upgrade our web page.

Pascal Batard has used Phycoerythrin and Allophycocyanin as a
donor/acceptor pair in FacsCalibur, and has shown that the sizes of the
fluorophores can decrease the maximum FRET efficiency obtainable. This
manuscript has been also accepted in the Cytometry and will be published
within a few month.  You can approach Pascal Batard for preprints.

Pascal Batard <Pascal.Batard@inst.hospvd.ch>


The Use of Phycoerythrin and Allophycocyanin for Fluorescence Resonance
Energy Transfer Analyzed by Flow Cytometry: Advantages and Limitations

Pascal Batard1,*, Janos Szollosi2, Immanuel Luescher3, Jean-Charles
Cerottini3, Rob MacDonald3, Pedro Romero1


Regards

Janos


At 08:52 AM 5/15/2002 -0400, you wrote:

>Greetings everyone:
>
>Has anyone had any success doing FRET on a single or dual laser FACS
>analyzer? [ie-Facscan or Facscalibur]
>
>Suggestions of FRET pairs and tips would be greatly appreciated.
>
>All my best,
>
>-Rich
>
>
>
>Richard Konz, Jr.
>Scientist
>Manager; Flow and Image Cytometry Facility
>Department of Musculoskeletal Sciences
>Wyeth
>200 Cambridge Park Drive
>Cambridge, MA  02140
>Office:  617.665.5522
>Lab:  617.665.5551
>rkonz@wyeth.com

Janos Szollosi
University of Debrecen
Medical and Health Science Center
Faculty of Medicine
Department of Biophysics and Cell Biology
P.O.Box 39, Nagyerdei krt. 98
H-4012 Debrecen
Hungary
Phone/Fax:  (36) (52) 412-623
E-mail:  szollo@jaguar.dote.hu



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:26:10 EST