Kevyn Adler,

One of our clients wanted us to conduct endotoxin-free sorts on a FACS Vantage.
Our procedure entailed preparing the cytometer with the following run
sequentially through both the sheath tank and the sample port: 1% ES 7X,
10% bleach, 70% ethanol, distilled water, and finally "blood bank" quality
saline. We received no indication from the client that endotoxins were ever
present in the sorted samples, although I do not know if the samples were
tested
for endotoxin.

After cleaning your FACScan, I would recommend installing a new sheath filter
before the final saline wash. We use saline from Bio Test Medical, available
from Fisher Scientific, catalog number MSC1000B.

Good luck,
Eric