Re: FACS Calibur

From: Ray Lannigan (lannigan@tritechinc.com)
Date: Wed Nov 07 2001 - 09:00:43 EST


Hi Carol,
    There is an alternative to the 96 well plate autosampler that BD offers.
It is called the Automatic Microsampler system and is made by Cytek
Development. If you would like more information please contact me.
    As far as sorting on the Facscaliber, recovery of and viability of
sorted cells can be an issue. The inherent problem with the mechanical
sorting technique used in the FacsCaliber, is fluid constantly flows to your
sort vessels, even when it is not trying to sort a cell in the sort gate.
After two hours of sorting, your 1000 - 2000 targeted cells could be in
approx. 300mL of fluid. The sorting mechanism is a catcher tube that goes
into the core stream of cells and attempts to catch the cell of interest.
When a cell is moving at a velocity of 6 meters/sec forcing a small tube
into its path, then forcing it to go through that tube can put relatively
extreme shear forces upon it. Stream-in-air with the electric charge sorting
is the way to go.
Ray


-----Original Message-----
From: Carol Mazurek <mazurek@mpi.com>
To: cyto-inbox
Date: Tuesday, November 06, 2001 6:44 PM
Subject: FACS Calibur


>
>I realize that the FACS Calibur multiwell autosampler is fairly new, but
>is anyone actively using one?  Could you please share your successes and
>failures with it?   Does it do what you need?  What would you change
>about it?  Would you recommend it to others?
>
>Is there anyone with experience at sorting cells using the FACS
>Calibur?   I realize it has limitations, and I'm not expecting FACS
>Vantage-quality sorting.   I have a cell line that is expressing a
>recombinant protein and I want to sort the highest protein expressors
>(e.g., top 1% or 10%) in that population.  What is the likelihood that I
>could get 1000 - 2000 targeted cells at >98% purity after sorting for an
>hour or two?
>
>
>Carol Mazurek
>Millennium Pharmaceuticals, Inc.
>mazurek@mpi.com
>