I disagree that: "The only argument is loss of fluorescence during the sort." If you have two populations that are close enough together the populations overlap. If there is, for instance, 50% overlap then the best you can hope for is a 50% pure population after sorting. I think this is very likely what is happening here. The real problem isn't the sorter, it's the prep. You need to get better separation between the neg and pos populations. There are amplification kits available. Here's another example to make my point. In another life I had a client that liked to sort a narrow band of a very broad GFP+ peak. Say we sorted a band that was between 800 and 1000 MFI. He expected a population upon reanalysis to show up "right between the goal posts" on a single color plot. It doesn't work that way. Upon reanalysis you get a normal Gaussian distribution and most of the cells are between the sort markers but you will get spillover on BOTH sides of the gate. Look at it this way: If you ran the SAME exact cell through the cytometer 10,000 times, you wouldn't get a line on the histogram with all events stacked in the same channel. You would get a normal Gaussian distribution predictable by the (if) known uncertainties in the measurement. Think of your calibration particles. They are very uniform in their spectral properties, but they don't all give the same exact fluorescence, although the width of the peak (CV) is much smaller because of this uniformity. One more thing and then I'll desist. Try seeing what happens if you sort half of a normal distribution. You might even compare the high half and the low half. Upon reanalysis these should look very much like what you started with before sorting. Oh, when I say normal, I mean that it can't be a broad, flat, 3 decade GFP-like peak which is really a string of overlapping peaks. David McFarland GlaxoSmithKline ----- Forwarded by David C McFarland/DEV/PHRD/SB_PLC on 02-Nov-2001 09:00 ----- Volker_Eckstein@med.uni-heidelberg.de 31-Oct-2001 11:10 To: cytometry cc: Subject: AW: Sort purity Hi Simon, I have had the same problem you described and I discussed it with BD's sorter specialists. Various points have to be taken into account: Weak fluorescence can result from low antigen density, bad protein/fluorochrom ratio or insufficient amount of used antibody. Even the antigen distribution or the fluorochromes distribution (patches) when labelling cell membranes can give slightly different signals when passing the analyzing point. The fluorochrome itself (tandemconjugates or not) can give weaker or stronger signals. If a sort gate is set all cells or particles which fit these criteria will be sorted if they can be sorted (no coincidence). When passing the laser light, weak fluorecent cells may loose fluorescence by bleaching. Therefore reanalyses often show the sorted cells not fit the sort gate again. And if the desired population was close to the negative unwanted ones it is possible that during reanalyses the fluorescence of the sorted cells fall into the negatives. The only argument is loss of fluorescence during the sort. RegardsVolker Volker Eckstein PhD Dept. of Internal Medicine V Medical School of the University University of Heidelberg Heidelberg - GERMANY volker_eckstein@med.uni-heidelberg.de -----Ursprüngliche Nachricht----- Von: Simon Monard [ mailto:smonard@trudeauinstitute.org <mailto:smonard@trudeauinstitute.org> ] Gesendet am: Montag, 29. Oktober 2001 20:22 An: Cytometry Mailing List Betreff: Sort purity Hi folks. Can anyone point me to a discussion of ways to estimate the purity of a sorted population of cells. This is clearly easy when you are sorting very bright cells from negatives but more problematic when sorting cells from a "shoulder". I remember seeing a nice discussion on this subject but cannot remember where. I've been sorting some very weakly positive cells from a population, these cells post sort overlap the negative cells quite a bit. My customer seems unhappy about that. Thanks Simon Monard FACS Lab Manager Trudeau Institute Saranac Lake NY12983 Ph 518 891 3080 X352
This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:37 EST