AW: Sort purity

From: Eckstein, Volker (Volker_Eckstein@med.uni-heidelberg.de)
Date: Wed Oct 31 2001 - 11:10:47 EST


Hi Simon,

I have had the same problem you described and I discussed it with BD's
sorter specialists.
Various points have to be taken into account:
Weak fluorescence can result from low antigen density, bad
protein/fluorochrom ratio or insufficient amount of used antibody. Even the
antigen distribution or the fluorochromes distribution (patches) when
labelling cell membranes can give slightly different signals when passing
the analyzing point. The fluorochrome itself (tandemconjugates or not) can
give weaker or stronger signals.
If a sort gate is set all cells or particles which fit these criteria will
be sorted if they can be sorted (no coincidence). When passing the laser
light, weak fluorecent cells may loose fluorescence by bleaching. Therefore
reanalyses often show the sorted cells not fit the sort gate again. And if
the desired population was close to the negative unwanted ones it is
possible that during reanalyses the fluorescence of the sorted cells fall
into the negatives.
The only argument is loss of fluorescence during the sort.

RegardsVolker

Volker Eckstein PhD
Dept. of Internal Medicine V
Medical School of the University
University of Heidelberg
Heidelberg - GERMANY
volker_eckstein@med.uni-heidelberg.de



-----Ursprüngliche Nachricht-----
Von: Simon Monard [ mailto:smonard@trudeauinstitute.org
<mailto:smonard@trudeauinstitute.org> ]
Gesendet am: Montag, 29. Oktober 2001 20:22
An: Cytometry Mailing List
Betreff: Sort purity


Hi folks.
Can anyone point me to a discussion of ways to estimate the purity of a
sorted population
of cells. This is clearly easy when you are sorting very bright cells from
negatives
but more problematic when sorting cells from a "shoulder". I remember seeing
a nice
discussion on this subject but cannot remember where. I've been sorting some
very
weakly positive cells from a population, these cells post sort overlap the
negative
cells quite a bit. My customer seems unhappy about  that.