Re: power meter

From: Emmanuel Gustin (Emmanuel.Gustin@Tibotec.be)
Date: Fri Jun 22 2001 - 02:36:38 EST


> Coherent does have an extensive line of sturdier and more expensive
> power meters.

So do several other manufacturers, so one can shop around. One
factor in the cost is the accuracy, you can get a cheap power
meter accurate only to about 10% (LaserCheck is specified for
5% accuracy) but you have to pay nearly $2000 for one that
is accurate to about 1%. And you really have to take good care
of them if you want to retain that accuracy!

There a two common types of power meters: Those that use a
photodiode to read the signal, and those to use the heating of a
black target surface to determine power. Photodiode power
meters react faster and more easily read low powers, but they
are strongly wavelength-dependent, so you have to dial in the
wavelength or use a calibration chart, and they have a more
limited spectral range. Thermal power meters have to be
adjusted for the environmental temperature if you want to read
off low powers (they will react to the temperature of your hand)
but they have a flat efficiency curve over a much wider range
(say 250nm to 11000nm).

Emmanuel Gustin



This archive was generated by hypermail 2b29 : Sun Jan 05 2003 - 19:01:23 EST