sorting binucleate cells

From: Michael Ormerod (Michael_Ormerod@compuserve.com)
Date: Sun Mar 04 2001 - 15:02:51 EST


You might be able to separate binucleate cells from two mononucleated cells
by looking at the DNA signal, showing peak height against area. This would
be best attempted on an instrument that has a narrow laser beam, basically
those that use crossed cylindrical lenses to focus the laser beam, such as
the Beckam Coulter instruments and some of thte Partec/Dako machines. You
would still have difficulty in distinguishing binucleates in G1 from
mononucleates in G2, unless you also labelled for cyclin B1.

If you have access to a Beckman Coulter Elite, you could also try looking
at time-of-flight on peak forward scatter. This will usually separate
single cells from doublets.

Michael Ormerod
34 Wray Park Road
Reigate RH2 ODE
Telephone: +44 (0)1737 241726
FAX: +44 (0)1737 226736
Mobile telephone: 07802 293242

Message text written by "Antony Bakke"
>
I have a PI who wants to sort binucleate from mononucleate myocytes.  The
problem is that
some of the mononucleated cells clump together during preparation.  Has
anyone sorted
binucleate cells from clumps of two mononucleated cells?  We only have
Hoechst staining
as a marker at present, since there are no known surface markers.  The
binucleates are
assumed to stop dividing and to only grow by enlarging.  The mononucleates
still divide.

Thank you,
Tony Bakke, PhD
Director Oregon Cancer Center Flow Cytometry Core<