Hi

1. The MACS beads are indeed tiny and do not interfere with sorting at all.

2. If you are checking the purity of your cells with an antibody recognizing the same
epitope as the MACS beads, you will have problems as many or most of the epitopes
will already be occupied with the antibodies attached to the beads. You need another
antibody which does not compete with the first one.

3. My Vantage is stable over many hours with only minor adjustments of the amplitude knob
to keep the phase and break-off in the same place. You may have to back-flush and perhaps
do a nozzle flush every now and then to clear gunk from the nozzle depending on sample
prep. If you are sorting rare cells one trick is to increase the sample rate up to about
the same as the drop drive frequency, change the theshold to the fluorescent channel
which you have the rare cell marker in, increase the theshold until all the negative
cells disappear and change sort mode to enrich. If your original sample rate was 27K
(with no turbo sort and 70u noz) and your population is 1% then after increasing the
theshold the new sample rate will be about 270. With a three drop envelope your purity
will be in the region of 33% if you sort all cells above theshold. You then spin then
down and resort using normal theshold, rate and normal-R sort mode. When doing your
enrichment sort you should make your sort windows generous, scatter will not be that
tight. You should also make sure you have no lumps in your samples, hanging a filter
off your sample probe is best.

Good luck



Simon Monard
FACS Lab Manager
Trudeau Institute
Saranac Lake
NY12983

Ph 518 891 3080 X352


>>> Carolyn Jefferiss <prscmj@bath.ac.uk> - 8/8/2000 4:35 AM >>>

Dear Everyone,
Three things:-
1) Magnetic beads and Flow Sorting:
I have presumed that the beds (Miltenyi) would interfere with sorting
by flow cytometry, even though they are absolutely titchy. Please say
if you have sorted cells from a population containing
magnetically-labelled cells, to good effect.

2) Magnetically labelled cells:
When checking magnetically labelled and isolated cells for purity by
flow cytometry, has anyone looked at loss of label? I was wondering
whether or not there is a significant loss of label somewhere from
the positive fraction in coming off the column or later. - I do
everything at 4degrees and despite using the usual concentrations of
conjugated second abs see a significant proportion of totally
unlabelled cells in the fraction which was positively bound to the
magnetic column.

3)Sorting time:
Finally, When sorting a rare population from whole bone marrow, what
kind of sort rate should be used? How stable could I expect "my"
Vantage to be over the number of hours I'd need to use it? I am
extremely impressed by the indication that other users seem to have a
stable settup for many hours. I have never sorted for more than four
hours. I'd need to run at least 100 million cells to get enough of my
chosen population.


Thank you for your attention.
Carolyn Jefferiss