FW: paraformaldehyde vs formaldehyde

From: DARZYNKIEWICZ ZBIGNIEW (DARZYNK@nymc.edu)
Date: Tue Aug 08 2000 - 15:01:20 EST


> -----Original Message-----
> From: DARZYNKIEWICZ ZBIGNIEW [SMTP:DARZYNK@nymc.edu]
> Sent: Wednesday, May 31, 2000 4:34 PM
> To:	Cytometry Mailing List
> Subject:	FW: paraformaldehyde vs formaldehyde
>
>
>
>
> > A year ago I responded to similar querry about paraformaldehyde vs
> > formaldehyde. I am enclosing my response below..
> >
> >
> > -----Original Message-----
> > From: DARZYNKIEWICZ ZBIGNIEW [SMTP:DARZYNK@nymc.edu]
> > Sent: Friday, May 21, 1999 2:52 PM
> > To: Cytometry Mailing List
> > Subject:	paraformaldehyde vs formaldehyde
> >
> >
> > Paraformaldehyde is a polymerized form formaldehyde. It is hardly
> soluble
> > and it cannot be used as a fixative. Only formaldehyde is used as a
> > fixative.  However, formaldehyde in aqueous solutions spontaneously
> > polymerizes. Therefore, methanol is often added to slowdown the
> > polymerization reaction. Solutions of formaldehyde (usually ~ 37%) in
> > water,
> > containig 10-15 % methanol as a preservative are generally called
> > "formaldehyde"; such solutions are being sold by most reagent companies.
> > Solutions further diluted (4-10 %) received name "formalin".
> > Methanol-free formaldehyde, which sometimes is preferred (e.g. for
> fixing
> > cells for some some histochemical  reactions or in immunocytochemistry),
> > can
> > be obtained by hydrolysis of paraformaldehyde. This is usually done by
> > extensive heating of paraformaldehyde solutions. Because of this
> procedure
> > the methanol-free formaldehyde received (incorrrectly) the name
> > "paraformaldehyde". In the past, this was the most common way to obtain
> > methanol-free formaldehyde. Unfortunately, this incorrect name  is still
> > often used in the literature, generating the confusion. The
> methanol-free
> > formaldehyde solutions can now be purchased. Some are called
> "ultrapure".
> > We
> > purchase such solutions (10%) from Polysciences, Inc. (800-523-2575);
> they
> > can be stored at room temperature. I would not recommend, however, to
> > store
> > them longer than one year, since formaldehyde in these solutions still
> has
> > tendency to polymerize. It should be noted that all formaldehyde
> solutions
> > are highly toxic and carcinogenic.
>
 Zbigniew Darzynkiewicz