Re: sorting potato protoplasts

From: Howard Shapiro (hms@shapirolab.com)
Date: Sat Jul 22 2000 - 20:29:14 EST


Joan Kalnitsky writes-


>         I have a client who would like to sort potato protoplasts.  We are
>wondering if anyone out there does this and if so if anyone has any input
>into whether Hoechts or DAPI would be a better choice.  Do both of these
>dyes work without affecting the viability of the cells.  They definately
>want live cells when we are done.
>         What about 7-AAD.  We currently run a lot of 7-AAD in our lab
> looking at
>apoptosis. I know it is used as a DNA dye.  Does it provide ploidy peaks?
>If so does anyone have any input as to whether this dye would be a good
>choice for this undertaking.
>         One other question while i'm at it.  For the DAPI and Hoechts, what
>filters would you recomend using to measure emission?  In the literature
>they had, there were conflicting reports of wavelengths.

Hoechst 33342 definitely gets into live mammalian cells better than either
DAPI or 7-AAD, to which the membranes of viable cells are normally
impermeable.  I would guess that your protoplasts would take up the Hoechst
dye; if they took up 7-AAD, I'd be concerned about their viability, and I
wouldn't expect them to take up DAPI, although some viable bacteria seem
to.  In fact, it might be advisable to run Hoechst 33342 with 7-AAD, and
only sort the cells which don't take up 7-AAD.

If you have more than 5 mW of UV excitation, it doesn't much matter what
filter you use for Hoechst dye.  I generally use a 450 nm/ 20-30 nm
bandwidth filter to keep the 488 nm in the second beam out of the Hoechst
channel.

-Howard



This archive was generated by hypermail 2b29 : Sat Mar 10 2001 - 19:31:26 EST