Re: ClonTech DS-Red on Vantage

From: Bill Telford (telfordw@box-t.nih.gov)
Date: Tue Jun 06 2000 - 15:34:08 EST


Hi Jim...

Teresa and Bob Hawley at the American Red Cross Holland Lab have been using
DsRed extensively and designing filter configurations for it (with and
without other FPs) on our Vantage SE at the NIH.  All of our collective
comments are below.

To begin, my commentary on the instrument end (with edits from Teresa)...

First, we can excite DsRed reasonably well with a 488 nm line (100 mW,
I-90) and detect it through a 585/42 or 610/20 filter.  We can also run
DsRed simultaneously with EGFP and EYFP using the standard Omega GFP/YFP
filter/dichroic set (525 LP splitter, 510/20 for GFP in FL1, 550/30 for YFP
in FL2) and a 560 SP and a 585/42 or 610/20 filter to split off and detect
the DsRed signal respectively (in the FL3 PMT).

Ar laser lines at 514.5 nm and Kr at 520 and 530 nm give improved DsRed
excitation.  However, the optimal excitation source for DsRed in our hands
is a yellow laser line - we use our I302 krypton laser tuned to 568 nm (35
mW), and detect DsRed through a 610/30 filter with a sharp cutoff at 595 nm
(Chroma) to reduce scattered laser light in the PMT.  The standard 610/20
filter BD supplies will also work with slightly higher noise.  The DsRed
signal is detected on the third signal path PMT (FL4).  Teresa has also
been doing 4-color analysis with GFP, YFP, CFP and DsRed using the argon
458 line as the primary laser as described by Beavis and Kalejta (Cytometry
37:68-73, 1999) and the Kr 568 nm line as the tertiary laser.  GFP and YFP
are collected as above - we use a 500 LP to split off the CFP signal,
collected though a 485/22 filter (FL3).   We have been able to analyze
GFP/YFP/CFP and DsRed this way simultaneously without 568 nm laser light
leaking into the YFP PMT (FL2), since the primary and tertiary beams and
signals are sufficiently separated -however, the instrument alignment has
to be perfect and is hard to reproduce from day to day.  We get more
consistent results with a 568 nm blocking filter in front of the YFP PMT
(FL2).  Considerable intra- and interbeam compensation (and teeth-gritting
instrument alignments) are required to run multiple FPs, but it can be done.

That's the technical end - here are Teresa and Bob's comments on
applications...

"DsRed's relatively low  extinction coefficient and quantum yield appear to
make it inappropriate for
the study of expression at the level of single gene copy in most instances.
However, we feel that in situations where high expression can be achieved
(due to a strong promoter or integration of multiple gene copies), DsRed's
red-shifted spectrum makes it rather attractive. At present, the best
application is in vivo tracking of stably marked cells (probably isolated by
sorting) which can be utilized in conjunction with similarly engineered
cells expressing the other FPs or surface-labeling antigens."

So, DsRed in its present incarnation is certainly possible to use on a
Vantage, but its extinction coefficient and yield are lower the GFP.
Nevertheless, its EX/EM characteristics are potentially very useful for
multiple FP experiments or immunophenotyping experiments.  I'm sure the
good folks at ClonTech are engineering improvements as we speak.

Bill Telford


 At 12:57 PM 5/31/00 -0500, Houston, Jim wrote:
>
>We are having some problems running this on the Vantage.
>Anybody trying this out that has had success?
>
>
>Jim Houston
>Coordinator, Flow Cytometry and Sorting
>BMT/CGT
>St. Jude Children's Research Hospital
>332 North Lauderdale
>Memphis, TN 38105
>
>ph:901-495-4155 or 2926
>fax:901-495-3966
>



This archive was generated by hypermail 2b29 : Sat Mar 10 2001 - 19:31:22 EST