Joseph,
The ECFP form will not work with your typical flow setup.  You'll
require a Krypton laser, with optics to provide a 407nm excitation
line.  Look at J. Dezz Ropp et.al. (Cytometry 21:309-317, 1995; Aequorea
green fluorescent protein analysis by flow cytometry) as a good early
reference describing these parameters.
Keep in mind, though, that this difficulty may be overcome by using the
recently introduced DsRed variant (Clontech).  While the peak excitation
is shown at 558nm, the claim is that it is also nicely excitable at
488nm.
MAK.

Joseph Webster wrote:

> I have a user who wishes to run Clontech EGFP and ECFP together; we
> can run EGFP O/K but we have not yet tried ECFP & I'm dubious....
>
> Has anyone used ECFP (cyan) in a flow cytometer?
> If so, what laser & wavelength excitation?
>
> The Clontech website says excitation is 433 with a secondary
> absorption peak at 453.
> No spectra shown, so I don't know how broad the absorption curve
> might be.
> We don't have any 433 line available, would the argon 457 line work?
>
> Any ideas &/or experience?
>
> Thanks, Joseph.
> --
> Joseph Webster, Flow Cytometry Facility
> Centenary Institute, Sydney AUSTRALIA.

--