Re: FITC PE 7AAD & APC on dual laser Calibur...

From: Bill Telford (telfordw@box-t.nih.gov)
Date: Tue Jan 18 2000 - 13:12:56 EST


Hi Stephen...

We routinely do FITC, PE and APC immunophenotyping with 7-AAD at 5 ug/ml as
a viability indicator.  We also have a cell death assay using Oncoimmunin's
PhiPhiLux G1D2 (which emits around the FITC range), a PE-conjugated
antibody, 7-AAD and APC-annexin V all simultaneously.  Unlike PE-Cy5, 7-AAD
has almost no crossbeam interference issues with APC.  We can run both
assays on either our Vantage with a HeNe laser or Calibur with the red
diode with no crossbeam compensation between 7-AAD and APC.  We DO,
however, notice some mild quenching of the PE signal with 7-AAD present,
particularly at higher 7-AAD concentrations (such as those needed for DNA
cell cycle).  Not a big problem, but you should watch out for it.  I don't
have a good explanation for why it happens (perhaps some sort of energy
transfer).

Hope this helps.

Bill

________________
Bill Telford, Ph.D.
NCI MB Flow Cytometry Core Laboratory
National Cancer Institute
National Institutes of Health
Bethesda, MD

At 06:30 PM 1/17/00 -0500, Fang-Yao (Stephen) Hou wrote:
>
>Dear FLOW users:
>
>I am wondering if anyone here has experience in running the 4-color panel,
>FITC PE 7AAD & APC on a dual-laser Calibur. Can we separate the signals
between
>7AAD and APC? I appreciate any experience or publication in this aspect.
Thanks
>
>Stephen Hou