Re: side population wtih LSR

From: Derek Davies (daviesd2@cancer.org.uk)
Date: Thu Feb 28 2002 - 05:08:57 EST


Hi Kimmo,

I think your problems are stemming (no pun intended, honest!) from the
filter set-up of the LSR. If you look at the laser paths, the UV signals
are directed to the FL4/5 PMTs via a 510LP filter so only fluorescence
below 510nm will get to the detectors. As you say, this is a bit of a
problem when looking for signals in the far red in the SP population.
The problem will be worse if you alter the filter in front of FL4 to say
a 650LP; keeping a 510/20 there will give you some signal but probably
not far enough out in the red for the 'classical' SP population.

There is no off the shelf solution as far as I can see other than
modifying the LSR considerably - I believe BD may be working on this. It
would be an adavnatge to be able to do SP analysis on a benchtop and I
belive that the Cytomation CyAn may be better equipped to do this -
currently we do all stem cell analysis on the MoFlo.

Derek

On Wed, 27 Feb 2002, Porkka Kimmo wrote:
> Has anyone have experience on running a side population analysis of human
> bone marrow with the BD LSR? I'm using the Hoecst 33342 dye and a FITC cell
> surface marker and following the published protocol to the letter, but still
> cannot see any Hoechst fluorescence on the red channel. Filter setup?
> Software compensation? Any help would be much appreciated.

************************************************************************
Derek Davies                       Voice: (44) 020 7269 3394
FACS Laboratory,                   FAX: (44) 020 7269 3100
Cancer Research UK,                e_mail:derek.davies@cancer.org.uk
London Research Institute,	   mobile: 07790 604112
44 Lincolns Inn Fields,
London, UK.

Web Page: http://sci.cancerresearchuk.org/axp/facs/davies/index.html

In tenebris lux
*************************************************************************



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:59:26 EST