Re: Antibody Conjugation Vendor and Alexa633

From: Mark Shlomchik (mark.shlomchik@yale.edu)
Date: Tue Feb 05 2002 - 12:09:05 EST


To all:

We have found that Alexa647 seems to match the HeNe laser the best,
with the best compromise of absorption and emission spectra, at least
for the FACScalibur filter set and probably for most sorters that are
set up for APC.  It certainly gives much better signal than 633,
conjugating the same protein preps.  While 633 gets excited more
optimally by the HeNe, its Stokes shift is small and much of its
emission is cut off by the typical filters used for APC/FL4.

Although 647 has almost the same spectral qualities as Cy5, for some
reason we seem to get superior reagents from the Alexa dye.  I would
also point out that (finally) Molecular Probes is now selling the
bulk hydroxysuccinimide ester reagent, making it much more economical
for those who want to do a number of conjugations.  However, none of
the Alexa dyes is going to be as bright as APC, so if brightness is
paramount, you will still want APC.

A question to Pete:  are you distinguishing the 630 and 660 signals
from the same excitation simultaneously?  If so, what filter sets do
you use and what are the compensations, etc.

Mark Shlomchik, MD, PhD
Associate Professor of Laboratory Medicine
and Immunobiology
Yale University School of Medicine

At 4:41 PM -0500 2/1/02, Howard T Petrie wrote:
>Having spent many years conjugating many antibodies to
>phycobiliproteins, I can understand why you wouldn't want to do it
>yourself (although the results are ALWAYS better).  For the last
>couple of years, however, we've been using Alexa633 instead, and I
>have to say that even using the straight-out-of-the-box protocol
>supplied by Molecular Probes, the results are quite good (if you have
>enough antibody to spare and want to test a couple of chrome:Ab
>ratios the results can be even better).  I highly recommend giving it
>a try, it takes about one hour of hands-on time and about three hours
>in total.
>
>I should mention that Alexa660 works beautifully as well for flow, in
>fact better than 633 (I'd love to hear somebody at either MP or one
>of the equipment manufacturers explain why this should be the case;
>neither the optics nor the spectra would predict it).  In fact after
>years as a dye laser loyalist, we now use HeNe exclusively, since we
>can simultaneously excite and detect A633 and A660 instead of APC and
>TR.  Given the utter simplicity of conjugation of the Alexa dyes,
>it's worth considering.
>
>>	To All:
>>
>>	I am in need of getting two monoclonal antibodies conjugated to APC.
>>	If you have had this done by some company/lab and liked the result
>>would you please let me know.
>>
>>	thanks
>>
>>	Jim Houston
>
>--
>
>
>Howard T. Petrie, Ph.D.
>Head, Laboratory of Developmental Immunology
>Memorial Sloan-Kettering Cancer Center
>Box 341, 1275 York Avenue
>New York, NY 10021
>phone (212)639-2149
>fax (212)794-4019

--
Mark Shlomchik, MD, PhD
Associate Professor of Laboratory Medicine
and Immunobiology
Yale University School of Medicine

203-688-2089
203-688-2748 (fax)
mark.shlomchik@yale.edu



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:59:23 EST