Re: Please help: which anti-CD10 clones work well on lymph nodes ?

From: Jáksó Pál (Pal.Jakso@aok.pte.hu)
Date: Mon Jan 28 2002 - 05:13:18 EST


Dear flowers,

First of all thanks for your many helpful replies. I will try to answer 
your questions as fully as possible.
1. Paula Fukushima wrote: Do you parallel test your lots?
I haven't  done paralel testing because  I have this problem with CD10 
only in lymph node GC and FCL but in B-ALL works perfectly and I didn't 
noticed any problem between different lots with these later cases.
2. Maryalice Stetler-Stevenson wrote: Do you use FITC or PE labeled 
CD10? This is important. FITC is very dim and,
although it is good enough for ALL, it is not good enough for follicular
lymphoma.
I use CD10 FITC but I have a question too. I know that FITC is dimmer 
than PE but have you tested CD10 with the two different flourescent dye 
or your statement - "it is good enough for ALL, it is not good enough 
for follicular" - is only an assumption ? Of course this may be the 
solution of my problem	and Frederic Preffer's anwer affirms this too.
3. Frederic Preffer wrote: I have no specific experience with the clones 
you utilize. However, i have
no problem detecting follicular lymphomas or GC cells with CD10PE 
available
from Becton Dickinson.
4. Mike Suter wrote: We have used Dako CD10 with good success for 
diagnosing follicular lymphoma from lymph nodes.  Blood neutrophils are 
positive and serve as a useful measure for validating new lot numbers of 
antibody.  You might want to verify reactivity of your CD10 with blood 
neutrophils by flow cytometry.
In my experience our DAKO CD10 FITC reacts always with mature 
neutrophils of blood and I also use them as internal positve control but 
in lymph node biopsies and fine needle aspirates I have not enough 
neutrophil for control purposes. Do you use FITC or PE labelled CD10 ? 
This would be very important because of the above mentioned advice of 
Maryalice Stetler-Stevenson.
5. Lerachmiel Daskal wrote: Were you doing three or four colors? On what 
were you gating?
I use three colors: CD10 FITC, CD19 PE and CD45 PE-Cy5 and I always gate 
on lymphoid cells (R2) using CD45/SSC gating (except in case of kappa 
FITC/lambda PE/CD19 PE-Cy5 where CD19/SSC is used) as the attached 
figure shows below. Before the CD45/SSC gating I use FCS/SSC gating for 
exlude the debris (R1).
You can also see the dim or partial CD10 positivity in two of our cases. 
Unfortunaly still in many instances we don't see any CD10 positivity 
although cytology an and immunocytochemistry give the obvious diagnosis 
of FCL !

FCM of FCL in bone marrow:
<File attached: FCL_in_BM.jpg>

FCM of FCL in lymph node fine needle aspirate:
<File attached: FCL_in_lymphnode.jpg>

6. Dirk Van Bockstaele wrote: I also find no agreement with the 
pathological diagnosis and I'm using the
CD10-FITC monoclonal of Becton Dickinson: so the BD monoclonal is no good
choice either!	 I would be interested in the suggestions that you 
receive:
could you forward them to me?
I think your reply also confirms what Maryalice Stetler-Stevenson 
suggested, namely we should try to switch to CD10 PE from CD10 FITC. As 
you could read in Frederic Preffer answer CD10 PE from BD works fine in 
their hands. However it's possible that you don't use the same clone 
because BD sells two clones with FITC (W8E7 and HI10A) while the HI10A 
is the only one labelled with PE.

Finally I've found and article in the latest American Journal of 
Clinical Pathology on the web by Yin Xu:
Assessment of CD10 in the Diagnosis of Small B-Cell Lymphomas A 
Multiparameter Flow Cytometric Study. This article is accesible for 
anybody for free on the AJCP website: http://www.ajcp.com/ and currently 
can be found in the previews section. In this paper they found that of 
the 58 FLs, 57 were positive (98 sensitivity!!!) and they used CD10 
(W8E7 clone) FITC (!!!) from BD. In their case repertoire they saw three 
CD10 expression pattern:
1. Uniform strong expression: more than 50% of the cases showed obvious 
CD10 expression
2. Uniform dim expession
3. Partial CD10 expression.

I'm very eager to read you future comments,


Pal Jakso

University of Pecs
Faculty of Medicine
Dept. of Pathology
7643, Hungary
12. Szigeti str.