 The emmision from Alexa 350 peaks at about 440nm thus you should be using
a 450BP25 or maybe a 450BP40 filter to capture light. Of course if you are
running 488 and 350 colinear you must filter out the 488nm light. I have
seen very good signals from this dye running a UV laser at 50-100mw. You
should be able to get at least a decade of separation from negative to
positive with CD4 antibody. I recommend using about 1ug of Alexa 350
secondary for 1-2x10E6 cells. Also be aware that you have to peak your
signal with UV beads or equivalent at the detector to capture optimal
fluorescence. Good luck.