Re: best 4-color combinations

From: Regina Harley (rharley@medicine.umaryland.edu)
Date: Wed Nov 08 2000 - 17:04:13 EST


Avis-

We routinely use a 5 color set-up on our Elite of FITC/PE/ECD/PerCP/APC
using the 488 and HeNe lasers, so it is possible to use
FITC/PE/PerCP/APC on the Elite as well as the Calibur.

Be aware that you can not use any of the PE-Cy5 conjugates (Cychrome,
etc) with the HeNe in conjunction with APC as you will get results that
are falsely double positive (i.e. anything that is positive for PE-Cy5
will also appear positive for APC, whether it really is or not) because
the HeNe will directly excite the Cy5 portion of the conjugate. I
suppose it may be possible to color compensate this out but I suspect
this is more of a nightmare than you want to deal with.

cheers-

Regina

>>> Tom Mc Closkey <thomasm@nshs.edu> 11/07/00 12:21PM >>>

Avis Peters wrote:
>
> We are looking into performing 4-color immunophenotyping analysis
> (leukemia/lymphoma) that must be compatable on a Coulter Elite and BD
Calibur.
> We want to use FITC,PE, and APC and are wondering which would be the
best 3rd
> color-PE/Cy7, Percp/Cy5.5, APC/Cy7, ECD, or other fluorochomes?

    It will be difficult to find get a 4 color combo that will work on
both
machines.  Thae Calibur is designed for FITC/PE/perCP/APC



Regina H. Harley, M.S.