RE: laser, polarized vs. random

From: Robert C. Leif, Ph.D. (rleif@rleif.com)
Date: Mon Aug 14 2000 - 19:19:27 EST


From: Bob Leif
To: cyto-inbox

The Coulter VCS, clinical technology uses about 45 degrees with an
unpolarized laser to produce excellent light scattering distributions of
white blood cells. Therefore, any problems with the suboptimal configuration
of 90 degrees are not based on science, but are the result of present
commercial practice.

-----Original Message-----
From: Howard Shapiro [mailto:hms@shapirolab.com]
Sent: Saturday, August 12, 2000 6:07 PM
To: cyto-inbox
Subject: Re: laser, polarized vs. random



Andrew Beernink writes-

>Can somebody tell me how plarization of lasers impacts flow?  Which is
>recommended?

For a comprehensive discussion of that topic - which should be required
reading for everybody on the Cytometry Mailing List - see Asbury, Uy, and
van den Engh, Cytometry, 2000, 40:88-101.

To quote from that paper, "With anisotropy, two instruments can yield
different results, and both can be right".  However, there is no reason for
widespread panic just yet; the flow cytometer manufacturers usually put the
lasers in the same way in most instruments, i.e., vertically polarized in
the direction of flow, because this orientation is necessary in order to
get detectable orthogonal scatter signals.  Differences in fluorescence
response would be most likely to be seen in instruments with substantially
different geometries in the fluorescence collection optics.  But you should
read the paper.

-Howard