Maciej,

While I can't offer a clinical perspective on your finding and its
significance, here is some background on CD11b/CD18:

CD18 can combine with the CD11a,b or c subunits to form to form the
integrins LFA-1, Mac-1 and p150,95. The CD11b/CD18 +/+ cells are
expressing Mac-1, but can also have the other integrins. The
CD18+/11b- cells are negative for Mac-1, but are expressing one or
more of the other CD18-utilizing integrins.

One of the ligands for Mac-1 is iC3b, a complement component which is
of course soluble; other soluble ligands are fibrinogen and factor X.
These could be one reason why staining whole blood is contraindicated.

Do you know whether the patient has a leukocyte adhesion defect? I
believe this is more likely to be seen with loss of CD18, where all 3
integrins are missing. CD11b is also highly expressed on CD8+
effector T lymphocytes (Hamann, D, J.Exp. Med., 1997); and you are
only seeing a decrease, not an absence of CD11b+ cells.

I'm sure there are other readers who can give you more in-depth info
on this, but its a start. For a reference try: Larson, RS and
Springer,TA (1990) Immunol. Rev. 114, 181-217.

good luck,

Mandy

>Greetings once again,
>
>We have recently received a blood specimen and were asked to stain
>for CD18/11b ( I love it how doctors don't explain the condition of
>the patient ahead of time, so we know what to look for )
>
>We found, that when compared to our NC's (N=300+ for this "tube" and
>growing) this patient had less +/+ lymphoctes (~30% vs. 12%). The
>expression of CD18 was not reduced - all of his leukocytes had >97%.
>
>A question came up: what is the significance of CD18+/CD11b-
>lymphocytes? what is the significance of a decrease in the +/+
>population?
>
>We do not routinely do CD11a or CD11c - are these mutually exclusive?
>I.e. it's either a, b or c but never any of them together?
>
>And finally, is the ligand for CD11b soluble and may have interfered
>with our staning?
>
>This staining was done in whole blood (which I just found is a
>no-no).
>
>I would appreciate if anyone could share any thoughts on this topic.
>
>Maciej
>
>__________________________________________________
>Do You Yahoo!?
>Send online invitations with Yahoo! Invites.
>http://invites.yahoo.com



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:55:45 EST