If anyone has used the mitochondrial membrane probes such as JC-1 or
MitoTracker can you tell me what works as a positive control?

Thanks.

Dear Butterfield,

At first, take care to use low probe conecntration, and this is the case
for all the probes.
inthe nM range will be the best.
The control is often to collapse to mitochondria membrane potential with
decoupling agents
which could be FCCP or mClCCP (sigma). You may realize a titration from 1
micromolar up to sometime 50 micromolar depending of the cell type and of
the medium used because the FSC in the medium or the others proteins will
trap part of the decoupling compounds.

See first Rottenberg et al. (1998) and our proper publication taking into
account that the amount of probe has to be reduced as much as possible up
to 1nM.

1.Rottenberg, H. & Wu, S. Quantitative assay by flow cytometry of the
mitochondrial membrane potential in intact cells. Biochim. Biophys. Acta
1404, 393-404. (1998).
2.Petit, P.X., O'Connor, J.E., Grunwald, D. & Brown, S.C. Analysis of the
membrane potential of rat- and mouse-liver mitochondria by flow cytometry
and possible applications. Eur. J. Biochem. 220, 389-397 (1990).
3.Zamzami, N. et al. Reduction in mitochondrial potential constitutes an
early irreversible step of programmed lymphocyte death in vivo. J. Exp.
Med. 181, 1661-1672 (1995).
4.Petit, P.X. et al. Alterations of mitochondrial structure and function
are early events of dexamethasone-induced thymocyte apoptosis. J. Cell.
Biol. 130, 157-167 (1995).

Sincerely yours

Patrice


                    \ | /
                    (o o)                           \ | /
 _______________oOo_____oOo_____oOo_  (o o)__oOo________________

Dr. Petit Patrice X.
Institut Cochin de Génétique Moléculaire
INSERM U129 - CHU Cochin Port-Royal
24, rue du Faubourg Saint-Jacques
F-75014 Paris, France.

Tel: 33 01 44 41 24 11
Fax: 33 01 44 41 24 21
E-mail: pxpetit@icgm.cochin.inserm.fr



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:55:45 EST