Re: CloneCyt/Single Cells

From: FACS_COPY@wehi.edu.au
Date: Tue Feb 01 2000 - 22:37:52 EST


Geoff,

If you can get 5 cells in each puddle on the lid every time, your
single cells are probably going into the well every time too.  To
confirm this by inverted microscope, don't forget to wait for your
cells to settle to the bottom; often they seem to sit at the edge of
the meniscus for a while before taking the plunge.  Round bottom wells
are good for locating the cells (see Boezeman et al Cytometry
28(4):305-10 1997).  If you prefer to see the single cells on a slide,
the fast evaporation of the droplet will give you problems but see
Alberti et al Cytometry 5(6):644-7 1984 and Kanz et al Cytometry
7(5):491-4 1986. 

Frank.