Re: CloneCyt/Single Cells

From: Dr Sasha Sreckovic (sasa.sreckovic@bris.ac.uk)
Date: Tue Feb 01 2000 - 18:19:54 EST


Dear Geoff,

When I was setting up my CloneCyt I was playing with fluorescent beads a
lot. If you can't take plate to the fluorescence microscope, put a slide
over it and then sort. Even when drop dries out you should see beads with no
problems. From my experience, Vantage is very reliable in sorting 1
particle. Take 3 drops envelope (effectively it's 4 for the CloneCyt, adds
1/2 drop both sides, just in case), go slow rate (if it's too fast you could
not have single target in 4 drops at all) and you should not miss out your
cell, although you are going to waste relatively lots of them while doing
this. I presume you did a few drop delay profiles before, that you can trust
to.

Hope this helps,
Sasha.

••••••••••••
Dr Sasha Sreckovic
Dept Path & Micro
University of Bristol
University Walk
Bristol, BS8 1TD, UK
Sasa.Sreckovic@bristol.ac.uk
0117-928-8606
••••••••••••

----------
>From: "geoff.morgan" <geoff.morgan@bbsrc.ac.uk>
>To: Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu>
>Subject: CloneCyt/Single Cells
>Date: Tue, Feb 1, 2000, 12:17 pm
>

>
> Dear Flowers,I have recently started using CloneCyt on a FacsVantage.We
> have benn sorting single cells into a 96 well plate which then has
> PCR. Cells are usually detected in half the wells.When I sort 5 cells
> onto a lid I can usually find 5 cells in most positions(by dissecting
> microscope), finding single cells is a lot harder.  I intend to try
m> again using blue stained cells.What sort of success do other people get
> any ideas how I can check/improve it   Thanks Geoff Morgan   Babraham,
> cambridge U.K.
>
> geoff.morgan@bbsrc.ac.uk
>
>
>
>
>
>
>
>
>
> 