From: David Chambers (davidc@ccmi.salk.edu)
Date: Fri Dec 03 1999 - 19:11:38 EST


We have an in-line hydrophobic filter (Whatman Vacu-Guard) between the vacuum
source and the waste tank on our Facstar+.  It works fine, however I should say
that we use house vacuum, rather than a pump.

-- David

On Thu, 02 Dec 1999, Art Roberts wrote:
> Just a thought regarding the damage to the pump caused by aspirating
> water, etc. into it:  Would it be possible to install a hydrophobic air
> filter in line between the waste receptacle and the pump in order to
> prevent the entry of damaging liquids without impeding the airflow?  I
> tried this on a vacuum gel drier and it seemed to work fine.  Has
> anyone tried it on a flow cytometer?
> 
> --Art
> 