Re: cell count

From: Ann Atzberger (Ann.Atzberger@EMBL-Heidelberg.de)
Date: Mon Dec 06 1999 - 06:02:07 EST


Hi Chris,

in our case the cells were actually lysing as they weren't compatible with
the medium being used. 
I'd also do a drop delay check. The cell number on the instrument counts
all the cells that fit in the sort gates but that does not mean it has
actually sorted it.

regards
Ann